UltiMaker uses functional, analytical and tracking cookies. Tracking cookies enhance your experience on our website and may also collect your personal data outside of Ultimaker websites. If you agree with the use of tracking cookies, click “I agree, continue browsing”. You can withdraw your consent at any time. If you do not consent with the use of tracking cookies, click “Refuse”. You can find more information about cookies on our Privacy and Cookie Policy page.
AD597 install and problem with Ultimaker 2+ thermocouple temperature
Posted
· AD597 install and problem with Ultimaker 2+ thermocouple temperature
Am not familiar with Marlin code, but as far as I can see, the AD597A uses a K-type thermocouple, and gives a reading of 1V at 100C, 2V at 200C etc up to 4V at 400C should be no problem, even up to 1200C providing the input supply voltage is at least 12V plus a bit. Farnell.com has a datasheet on the chip. This extract says
Since the output can only swing to within 2 V of the positive supply, the usable measurement temperature range will be restricted when positive supplies less than 15 V for the AD597...
. So if you give the chip a 5V supply, ti can only output at most a 3V signal corresponding to 300C. Suggest you check what voltage it's getting, and maybe give it 7V or even a bit more. Also make sure you're using a K-type thermocouple, not some thermistor or other J-ytpe or whatever TC.
I replaced a defective AD597A on my UMO, and in the process found that mine gets a 5V supply voltage. Enough for my ABS at 260C.
Link to post
Share on other sites
Create an account or sign in to comment
You need to be a member in order to leave a comment
Create an account
Sign up for a new account in our community. It's easy!
Here comes Cura 5.9 and in this stable release we have lots of material and printer profiles for UltiMaker printers, including the newly released Sketch Sprint. Additionally, scarf seams have been introduced alongside even more print settings and improvements. Check out the rest of this article to find out the details on all of that and more
We are happy to announce the next evolution in the UltiMaker 3D printer lineup: the UltiMaker Factor 4 industrial-grade 3D printer, designed to take manufacturing to new levels of efficiency and reliability. Factor 4 is an end-to-end 3D printing solution for light industrial applications
Recommended Posts
andywalter 7
Am not familiar with Marlin code, but as far as I can see, the AD597A uses a K-type thermocouple, and gives a reading of 1V at 100C, 2V at 200C etc up to 4V at 400C should be no problem, even up to 1200C providing the input supply voltage is at least 12V plus a bit. Farnell.com has a datasheet on the chip. This extract says
. So if you give the chip a 5V supply, ti can only output at most a 3V signal corresponding to 300C. Suggest you check what voltage it's getting, and maybe give it 7V or even a bit more. Also make sure you're using a K-type thermocouple, not some thermistor or other J-ytpe or whatever TC.
I replaced a defective AD597A on my UMO, and in the process found that mine gets a 5V supply voltage. Enough for my ABS at 260C.